Hi everyone,
We are a tinnitus research lab that started EEG research for the first time this year, so we are still slowly learning our way through building a pipeline and running recordings — and we would really appreciate the community’s insight on one recurring issue.
As shown in the attached screenshot, we frequently observe slow, rolling waveforms that appear across many channels at the same time. This pattern shows up in recordings from many different participants, not just one or two.
Our question is: what is the likely cause of this pattern? Specifically, we would like to know whether this looks like an equipment problem (something we could fix by changing our setup or having the amplifier serviced), or a subject-related issue such as drowsiness or reduced vigilance (something we could address by instructing and monitoring participants during the recording).
Our setup: Neuroscan Quik-Cap Net 64 channels, SynAmps RT amplifier in DC mode, sampling rate 1,000 Hz, 5-minute resting-state recordings. The slow waves sometimes appear mildly early in the recording, but they become clearly more frequent and larger in the later part, typically after about 2 minutes.
A side note on the screenshot: a few channels show large spiky deflections — this is a known issue with our cap (faulty electrodes), not the pattern we are asking about. We plan to remove and interpolate those channels during preprocessing, and the cap itself will be sent for repair soon. Our question is specifically about the slow, widespread rolling waves.
If you have any idea what might be going on, it would be a huge help for us.
Thank you very much!
Could you state the acquisition reference, display reference and filters used here? Given the faulty channels, one processing check is whether they were excluded from any average-reference calculation. Interpolating them afterwards doesn’t by itself remove their contribution from the other channels. MNE’s reference tutorial explains that exclusion. I’d establish the processing history before treating the widespread waveform as evidence of reduced vigilance; the screenshot alone doesn’t distinguish that from recording artefacts.
1 Like
Thank you very much for your thoughtful reply, and I apologize that my original explanation was not detailed enough.
To clarify our setup: we record the data using Curry 9, and during acquisition the data are recorded against the dedicated online reference electrode (REF) of the Quik-Cap. When we save the data in Curry 9 afterwards, we re-reference them to Cz. The common average reference is applied only later, as part of our analysis in MATLAB. The data in the screenshot are therefore still in the Cz reference, with only a 0.5–100 Hz band-pass filter and a 58–62 Hz notch filter applied.
I should also mention that the faulty electrodes on our cap are a recent problem. The slow waveforms spanning multiple channels, as shown in the screenshot, were already appearing frequently well before any electrodes failed.
To be honest, I am genuinely puzzled. Since the pattern was present before the electrode failures, I do not think it is caused by those particular electrodes. And even if they played some role, the average reference is only applied at the analysis stage after recording, so I suspect the pattern is not the result of contamination through the reference calculation either.
As we are still learning, there is a lot we are struggling with. I will carefully read the reference tutorial you kindly shared. Thank you again for taking the time to help us.
Thanks for clarifying. Since this screenshot precedes the common-average step, that later step can’t explain the pattern shown. If the original REF-referenced data are still available, could you compare the same segment against REF and against Cz, keeping filters and scale fixed? That would help check whether the widespread deflections are introduced by the Cz re-reference or already present in the acquisition.
1 Like
Thank you so much for your detailed and helpful reply.
I also suspect that the cause is most likely either a reference-related issue or the participants gradually becoming drowsy as the recording goes on. However, as I am still quite inexperienced, I had not fully realized that drowsiness can commonly appear as slow, rolling waves across many channels. Your explanation about the timing was truly eye-opening.
Following your suggestions, I will check for slow drift and sweat artifacts related to the DC mode, and cross-check the pattern with the EOG channels. I will also try brief verbal check-ins during the recording to see whether the pattern changes with vigilance.
I sincerely appreciate you taking the time to share your insight.
The screenshot looks a like a movement artefact to me. If you have a video recording you, you could check it there. Alternatively, you could record from a volunteer and ask them to produce different artefacts. That would give you a good idea of what different artefacts look like with your EEG setup.
1 Like
Thank you very much for sharing your opinion.
As we are an auditory (tinnitus) research lab, we record EEG inside a sound-attenuated booth, and the researcher can observe the participant from outside the booth throughout the recording. Each time these slow, rolling waves appeared, we checked on the participant, but we did not notice any particular movement.
That said, recording from a volunteer and asking them to produce different types of artifacts under various conditions, in order to see what each artifact looks like with our setup, is an approach I had never thought of. Thank you for this great suggestion. We will definitely give it a try.
For the volunteer test, I’d mark each action in the recording and leave quiet intervals between actions. Keeping filters and display scale fixed would make the comparison easier to interpret than matching waveform shapes from separate screenshots.