Uneven T1-w Brightness & fMRIPrep / FreeSurfer Failures After Scanner Upgrade

Dear Community,

In our latest data collection wave (Batch 4), T1-weighted images look darker and the intensity is inhomogeneous once converted to NIfTI, despite appearing almost normal in the raw DICOM viewer. Earlier batches from the same scanner (Batch 1-3) do not show this problem.

Our scanner is a Siemens 3 T that was upgraded in 2018 from a single-channel CP head coil (pre-XA software) to the current XA30 platform with a 64-channel head/neck array. A quick pilot scan revealed that enabling “Prescan Normalize” in the vendor’s Resolution → Filter menu restores normal contrast, suggesting the issue stems from that setting being OFF for Batch 4.

Key Symptoms from MRI quality checking

  • fMRIPrep 23.0.0 – recon-all aborts at Talairach (talairach_afd p < 0.005) → ~10 subjects
  • FreeSurfer 7.3.x – watershed error (“few WM voxels value 110”) → several
  • CAT12 – low QC grades even after bias-field correction → many
  • FSL VBM – GM segmentation fails (overall low intensity) → many
  • AFNI QC – SNR numerically OK, but anatomy visibly dark compared to other protocols → most

Visual comparison (see attachments):

  • “Problem” T1 NIfTI vs. “Typical” T1 from earlier batches
  • Raw DICOM screenshots – slight but noticeable contrast loss

Acquisition Details

  • 3T Siemens TIM Trio scanner and a 64-channel phased-array head coil
  • software version updated in 2018
  • 1 mm isotropic MPRAGE-type sequence
  • Coil-combine method: Adaptive
  • Prescan Normalize was OFF for Batch 4, ON for previous batches
  • Conversion: dcm2niix (2024-02-02) for Batch 4; earlier batches converted with a 2023 build.

Troubleshooting Performed

  • Bias-field correction (ANTS N4 & FSL restore) – modest brightening; However, the brain segmentation didn’t work well and FreeSurfer still produced extremely low or high values for subcortical volume, cortical thickness, and surface area in problematic scans.
  • Re-conversion with alternate dcm2niix options – no change.
  • New pilot scan with Prescan Normalize ON – intensity normal; processes fine.
  • Running fMRIPrep without surfaces (–fs-no-reconall) – works, but anatomical still dark and affects EPI→T1 co-registration quality.

Questions

  1. Is there any post-hoc method to replicate the vendor’s Prescan Normalize on existing NIfTIs?
  2. How much residual bias is acceptable before it starts impacting our functional data? We would be grateful for recommendations for co-registration settings in fMRIprep or other software.
  3. Recommended workflows to rescue these scans (stronger N4? SynthStrip uniformization? mri_robust_template for Talairach?)
  4. Practical exclusion criteria – e.g., CAT12 QC threshold or WM/GM contrast ratio – that others use to decide when to drop subjects vs. keep them.
  5. If we keep them, is it safer to skip surface-based analyses?

Any guidance on automated fixes, quality thresholds, or decision rules would be greatly appreciated.

Thank you!